Physio-Control, Inc.
11811 Willows Road NE
Redmond, WA 98052
800 442 1142
www.physio-control.com
Physio-Control is a division of Medtronic
February 21, 2007
Dear LIFEPAK® defibrillator/monitor & AED Customers:
Starting on March 11, 2007 the United States will begin a new schedule for Daylight Saving Time
(DST), as directed by the Energy Policy Act of 2005. With this change, DST will last one month longer
than previous years. In 2007, we will "Spring Forward" on March 11 and "Fall Back" on November 4.
The new DST schedule lasts from the second Sunday in March to the first Sunday in November. The
exact time change remains at 2:00am.
How Does this Affect LIFEPAK® Defibrillators?
The US Department of Energy is required to study the impacts of this change and report back to
Congress, which reserves the right to change things back. Because this change may only be
temporary, we will not be updating our devices at this time. If the changes do become permanent, we
will take appropriate action.
Physio-Control data management products such as CODE-STAT™ Data Review Software and DT
EXPRESS Data Transfer Software offer LIFEPAK devices the ability to correct time set on the devices.
Time on the LIFEPAK devices will adjust according to the time specified on the PC. Time
synchronization occurs when post-event data is downloaded to the data management application
running on the PC. This feature must be enabled for time synchronization to occur. Keep in mind the
PC clock must be adjusted prior to using the time synchronization feature. For details on how to
prepare your PC clock for the new Daylight Savings Time schedule, please visit
http://www.microsoft.com/office/timezone.mspx.
Device Specific Information:
In order to be consistent with the new DST schedule, the DST set-up option needs to be
turned off and the device time needs to be reset twice a year. We recommend you perform
the following steps for each LIFEPAK device you own:
LIFEPAK 12 defibrillator/monitor
The device clock will not adjust to this new DST change, so users will need to manually reset the
device clock and then turn off the Daylight Savings feature.
• The device clock can be adjusted by pressing “ON”, while holding down “OPTIONS” and “EVENT” to
enter the “SETUP MENU” (users should continue to hold down these controls until the passcode
screen appears).
• Enter passcode; “SETUP MENU” should appear
• Select “CLOCK” from the menu, then select “DATE/TIME” to enter the appropriate time
• Select “DST” and select “OFF”
LIFEPAK 20 defibrillator/monitor
The device clock will not adjust for this new DST change, so users will need to manually reset the
device clock and then turn off the Daylight Savings feature.
• Open the door on the front panel by pressing the “MANUAL” button
• The device clock can be adjusted by pressing “ON”, while holding down “OPTIONS” and “EVENT” to
enter the “SETUP MENU” (users should continue to hold down these controls until the passcode
screen appears).
• Enter passcode; “SETUP MENU” should appear
• Select “CLOCK” from the menu, then select “DATE/TIME” to enter the appropriate time
• Within the “CLOCK” menu, enter “DST” and select “OFF”
LIFEPAK 500 & LIFEPAK 500 DPS AEDs
• To change the date and time: Turn on the AED. (Be sure the AED has been off for at least 60
seconds and that nothing is connected to the AED.)
• Press and hold the ▲ or ► button for approximately three seconds until the AED displays the
date and time setting. A value blinking on and off indicates that the value can be changed. The
day, month, year, hour, and minutes values can be increased. The seconds value can be reset to
zero.
• To set the hour and minute, press the ▲ button to increase the value and the ► button to
advance to the next field.
• To reset the seconds value to zero, press the ▲ button once.
• After the time is set, press ON/OFF to turn off the AED.
LIFEPAK 1000 AED
• The device clock can be adjusted by entering the setup mode. Press and hold the two softkeys
located directly under the screen, then press the ON/OFF button. Use the softkeys to navigate and
make selections on the Setup screen.
• Select “General” and press “SELECT.”
• Select “DATE/TIME” by pressing “NEXT” to advance through the menu options.
• To enter the appropriate time, use the softkeys and follow the instructions on the screen.
• To exit setup mode, turn the defibrillator off. The changes are saved and will appear the next time
you turn the defibrillator on.
LIFEPAK CR® Plus and LIFEPAK EXPRESS® AEDs
The device clocks will not adjust for this new DST change. Users cannot manually reset the device
clock. However, each time you use the defibrillator, it digitally saves data, including date and time,
that can be transferred to a PC. If you choose to download the data and transfer to a data
management system, such as CODE-STAT™ Data Review Software and DT EXPRESS™ Data Transfer
Software, the time will adjust according to the time specified on the PC. Please follow the instructions
under Data Management to set your PC clock for the new DST change.
DATA MANAGEMENT PRODUCTS
For LIFENET® RS Receiving Station Customers, the new Daylight Savings Time schedule will also affect
your Dell PC clock. Please contact Technical Support to obtain information on which Microsoft OS patch
for Daylight Savings Time changes applies to your LIFENET RS system.
For CODE-STAT software Customers, the time synchronization feature must be enabled to adjust time
on LIFEPAK devices. The time synchronization feature can be enabled in the Device Data Reception
tab of the LIFENET System Configuration of the application.
For DT EXPRESS software Customers, the time synchronization feature must be enabled to adjust time
on LIFEPAK devices. The time synchronization feature can be enabled in the Communications tab of
the Configuration Properties of the application.
If you have difficulties with these instructions or need additional information please refer to the device
Operating Instructions or contact Physio-Control Technical Support at 1-800-442-1142.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: /pipermail/csemsmail_csems.vaems.org/attachments/20070227/79272ea2/attachment-0001.html
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 20070227_physio_dst.pdf
Type: application/pdf
Size: 67453 bytes
Desc: not available
Url : /pipermail/csemsmail_csems.vaems.org/attachments/20070227/79272ea2/attachment-0001.pdf
More information about the CSEMSmail
mailing list